  • Oxygen concentrator power requirements.

    I just un-plugged my father's oxygen concentrator from the Kill-O-Watt. 4.42 kwh for a 12 hour period. The thing runs 24/7 so 8.82 for a 24 hour period. So far in searching the internet for specifications I have found that manufacturers don't advertise much in the way of power requirements. Hoping someone on the forum may have already researched this topic, and could provide some guidance.
    Outback GVFX3648
    MS Classic 150 with wizzbang,
    Outback PSX-240 Autotransformer
    MS Combiner box, and E-Panel
    3500 watts of PV
    464ah GC2 battery bank @ 48 volts
    Honda EM2500 Gen Set.
    .

  • #2

    Re: Oxygen concentrator power requirements.

    Re: Oxygen concentrator power requirements.

    What sort of guidance? One that doesn't use such a whopping amount of power? That would be nice. That really is a lot of power use daily; no wonder they don't want to talk about it.

    Comment


    • #3

      Re: Oxygen concentrator power requirements.

      Re: Oxygen concentrator power requirements.

      Wow, that's a pretty heavy consumption, roughly 370 watts continuous. That's major for an off grid system. Over 8.8 Kwh/24 hours, that's enough to power perhaps 9 new full size Energy Star fridges! That sucks!
      1900 watts PV, (1000 watts PV feeding MidNite Classic 150; 900 watts PV + 160 watts micro hydro both feeding into a single shared Morningstar TS-MPPT-60) ; Xantrex Pure Sine 1800/12 for heavy loads; Xantrex Pure Sine 1000/12 on 24/7 for everything else; six Rolls Surrette 2 volt L16 @ 12 volts.
      Domestic hot water totally provided by the sun 8 months out of every year via thermal panel.

      Comment


      • #4

        Re: Oxygen concentrator power requirements.

        Re: Oxygen concentrator power requirements.

        Since the units are basically air compressors you can see why it would be a big power user. Hard to compress any significant volume of air without using a lot of power, and then having to run it all the time ...

        I wonder if any supplier of these would mind if you brought along the Kill-A-Watt and measured a few in operation? If they were smart they'd see it as a selling feature; "this one uses the least energy for the volume of oxygen produced".

        Comment


        • #5

          Re: Oxygen concentrator power requirements.

          Re: Oxygen concentrator power requirements.

          Interesting info on how it works. Looks like considering what it does/how it operates, perhaps it's no wonder it sucks back on the power.
          http://en.wikipedia.org/wiki/Oxygen_...entrators_work
          1900 watts PV, (1000 watts PV feeding MidNite Classic 150; 900 watts PV + 160 watts micro hydro both feeding into a single shared Morningstar TS-MPPT-60) ; Xantrex Pure Sine 1800/12 for heavy loads; Xantrex Pure Sine 1000/12 on 24/7 for everything else; six Rolls Surrette 2 volt L16 @ 12 volts.
          Domestic hot water totally provided by the sun 8 months out of every year via thermal panel.

          Comment


          • #6

            Re: Oxygen concentrator power requirements.

            Re: Oxygen concentrator power requirements.

            Originally posted by Cariboocoot View Post
            What sort of guidance? One that doesn't use such a whopping amount of power? That would be nice. That really is a lot of power use daily; no wonder they don't want to talk about it.
            There are some portable units out there. This thing is a dinosaur, and a rental at that. Maybe there has been some progress. I'll call up some medical supply stores and see if they have some info.
            Outback GVFX3648
            MS Classic 150 with wizzbang,
            Outback PSX-240 Autotransformer
            MS Combiner box, and E-Panel
            3500 watts of PV
            464ah GC2 battery bank @ 48 volts
            Honda EM2500 Gen Set.
            .

            Comment


            • #7

              Re: Oxygen concentrator power requirements.

              Re: Oxygen concentrator power requirements.

              Originally posted by waynefromnscanada View Post
              Interesting info on how it works. Looks like considering what it does/how it operates, perhaps it's no wonder it sucks back on the power.
              http://en.wikipedia.org/wiki/Oxygen_...entrators_work
              Yeah. It has a compressor that doesn't ever shut down.
              Outback GVFX3648
              MS Classic 150 with wizzbang,
              Outback PSX-240 Autotransformer
              MS Combiner box, and E-Panel
              3500 watts of PV
              464ah GC2 battery bank @ 48 volts
              Honda EM2500 Gen Set.
              .

              Comment


              • #8

                Re: Oxygen concentrator power requirements.

                Re: Oxygen concentrator power requirements.

                Some discussion on a CPAP that may give you some ideas,

                http://forum.solar-electric.com/show...highlight=cpap
                http://forum.solar-electric.com/show...highlight=cpap

                another about a feeding pump

                http://forum.solar-electric.com/show...ghlight=oxygen

                hope they help
                100% Off Grid @ 51* 46' N lat 124* 44' W long

                New House system: coming - 2 arrays @ 2240W
                #1) CL150 & Epanel 12V PVs 4S @140 W , 24V C&D AT-15P AGM @ 950Ah, Linksys wet54g - WiFi bridge, Cotek ST1500W inverter, TBS Omni-charger 30a-24v, ,
                #2) CL150 12V PVs 4S@140W , 24V C&D AT-15P AGM @ 950Ah
                to ADD: E-Panel and inverter
                Honda Eu3000is,Eu1000is

                Guest cabin system: 3 - 70W panels to BS 2000e CC, with 2 - 100 ah 12v SAFT wet NiCd's , 600W TSW Inverter

                Comment


                • #9

                  Re: Oxygen concentrator power requirements.

                  Re: Oxygen concentrator power requirements.

                  CPAP's are tiny compared to O2 concentrators. No comparison at all, other than they both move air somewhat.
                  http://tinyurl.com/LMR-Solar
                  http://tinyurl.com/LMR-BigLug
                  http://tinyurl.com/LMR-NiFe

                  Grundfos 10 SO5-9 with 3 wire Franklin motor (1/2hp 240V 1ph # 214505 ) on a timer for 3 hr noontime run - Runs off PV
                  Powerfab poletop PV mount | Listeroid 6/1 w/st5 gen head | XW6048 inverter/chgr | Iota 48V/15A charger | Morningstar 60A MPPT | 48V, 800A NiFe battery | 15, Evergreen 205w "12V" PV | Midnight ePanel || VEC1093 12V Charger | Maha C401 aa/aaa Charger | SureSine | Sunsaver MPPT

                  Comment


                  • #10

                    Re: Oxygen concentrator power requirements.

                    Re: Oxygen concentrator power requirements.

                    you haven't mentioned what you have in mind with this info. are you on grid and just want backup power for it in case of an outage and if so then for how long? if you are looking at off the grid then as was said this is a huge draw of power constantly running and could compromise an offgrid pv system from supplying enough power for that and other items if not accounted for in the initial design.
                    NIEL

                    Comment


                    • #11

                      Re: Oxygen concentrator power requirements.

                      Re: Oxygen concentrator power requirements.

                      For now I have grid power. So yes this is back-up, and I do have a generator. My current plan is for 8 to 10 280 watt panels facing due south. My shed is still under construction. I will be using a generator transfer switch panel to power selected circuits. The bottom line on the oxygen concentrator is to find a more efficient one. There are portable battery powered units available that can be operated on AC. I don't have any specific power consumption numbers on those yet. I hope to find on line operator manuals to get an idea. The portable units are expensive. The grid is pretty reliable here in southern KY. A typical outage would be from thunder storms or in the winter an occasional ice storm. We lost the grid a couple of years ago for 18 days due to ice.
                      Outback GVFX3648
                      MS Classic 150 with wizzbang,
                      Outback PSX-240 Autotransformer
                      MS Combiner box, and E-Panel
                      3500 watts of PV
                      464ah GC2 battery bank @ 48 volts
                      Honda EM2500 Gen Set.
                      .

                      Comment


                      • #12

                        Re: Oxygen concentrator power requirements.

                        Re: Oxygen concentrator power requirements.

                        I wonder if a couple days of O2 in tanks would be a better backup, than generators & fuel? If you have bad storms, solar is not effective if skys are cloudy or panels smashed up from falling ice.
                        http://tinyurl.com/LMR-Solar
                        http://tinyurl.com/LMR-BigLug
                        http://tinyurl.com/LMR-NiFe

                        Grundfos 10 SO5-9 with 3 wire Franklin motor (1/2hp 240V 1ph # 214505 ) on a timer for 3 hr noontime run - Runs off PV
                        Powerfab poletop PV mount | Listeroid 6/1 w/st5 gen head | XW6048 inverter/chgr | Iota 48V/15A charger | Morningstar 60A MPPT | 48V, 800A NiFe battery | 15, Evergreen 205w "12V" PV | Midnight ePanel || VEC1093 12V Charger | Maha C401 aa/aaa Charger | SureSine | Sunsaver MPPT

                        Comment


                        • #13

                          Re: Oxygen concentrator power requirements.

                          Re: Oxygen concentrator power requirements.

                          Originally posted by mike95490 View Post
                          I wonder if a couple days of O2 in tanks would be a better backup, than generators & fuel? If you have bad storms, solar is not effective if skys are cloudy or panels smashed up from falling ice.
                          We have O2 tanks. Ice storms are not the same as hail storms. An ice storm is the result of freezing rain. I read somewhere that a Midnight Solar Classic charge controller can be told to melt snow or ice off of the PV array. (I may have dreamed that)
                          Outback GVFX3648
                          MS Classic 150 with wizzbang,
                          Outback PSX-240 Autotransformer
                          MS Combiner box, and E-Panel
                          3500 watts of PV
                          464ah GC2 battery bank @ 48 volts
                          Honda EM2500 Gen Set.
                          .

                          Comment


                          • #14

                            Re: Oxygen concentrator power requirements.

                            Re: Oxygen concentrator power requirements.

                            I think there was talk about melting ice off of solar panels--But, so far, that has never been implemented (as far as I know--which is not very far ).

                            -Bill
                            20x BP 4175B panels (replacement) + Xantrex GT 3.3 inverter for 3kW Grid Tied system + Honda eu2000i Inverter/Generator for emergency backup.

                            Comment


                            • #15

                              Re: Oxygen concentrator power requirements.

                              Re: Oxygen concentrator power requirements.

                              Originally posted by BB. View Post
                              I think there was talk about melting ice off of solar panels--But, so far, that has never been implemented (as far as I know--which is not very far ).

                              -Bill
                              Check out this link. http://www.wholesalesolar.com/produc...ontroller.html Here is a copy / paste of the statement.
                              "Another significant advance with the Classic is the ability to access your power system remotely via the internet. If you’re out of town, you can call in to the system to troubleshoot problems, and get it back in working order. If you have plenty of excess power, you can even turn on the snow-melting feature while you’re away. Remote control of Midnite power systems will be invaluable for installers, dealers and homeowners alike. "
                              Outback GVFX3648
                              MS Classic 150 with wizzbang,
                              Outback PSX-240 Autotransformer
                              MS Combiner box, and E-Panel
                              3500 watts of PV
                              464ah GC2 battery bank @ 48 volts
                              Honda EM2500 Gen Set.
                              .